Understanding differences in analog noise between PCIe-6255 and USB-6001

My team is using a DAQ to read a differential analog measurement from a strain gauge amplifier. Initially, we had been using a USB-6001 DAQ to test the system wiring and debug. However, for the actual application, we plan to use a PCIe-6255 installed in a PC with an SCB-68A terminal block to sample the differential signal as well as several other sensors in real-time. In transitioning over to using the PCIe-6255, we noticed that, generally, noise in our signal is more pronounced for the PCIe-6255 than the USB-6001.

 

Attached is some example data from the strain gauge collected separately at 1000 Hz using each device. The strain gauge bridge is installed on a model turbine blade that is being operated by a servomotor, so oscillatory readings are expected. The underlying signal for the USB-6001 is distinct, outside of semi-frequent larger spikes which could be easily removed with a software filter. Conversely, for the PCIe-6255 the signal is overall fuzzier. In both cases, everything upstream of the DAQs is identical: the DAQs are plugged into the same computer in the same environment, with terminal blocks located in roughly the same physical location. Cabling and shielding of the analog signal is identical between the two cases.

 

PCIe USB comparison.png

 

The main difference between these two cases is simply changing is the DAQ that the differential wires are plugged into. However, there are a few other differences at the DAQs that I should note:

  • I did find that I needed to include input bias resistors (per this guide) between the analog +/- inputs and analog ground for the PCIe-6255 in order to avoid saturated readings, whereas these resistors were not necessary for the USB-6001 (i.e., it was adequate to simply connect the inputs to AI +/-).
  • For the PCIe-6255, other instruments (servomotor encoder, 6-axis load cells) are also wired into the SCB-68A, although I observed similar differences in noise characteristics when these instruments were powered off. There is electrical isolation between the servomotor and the SCB-68A.

What might be the mechanisms for different noise characteristics between these two devices, and how might I reduce the overall noise of the PCIe-6255? Why are input bias resistors necessary for the PCIe-6255, but not for the USB-6001? Is the USB-6001 performing some sort of filtering that the PCIe-6255 is not, or perhaps the PCIe-6255 is more sensitive?

0 Kudos
Message 1 of 2
(65 Views)

The USB-6001 is an entry-level DAQ and USB-powered; this means that if you're running the DAQ off a laptop on its battery, it is essentially isolated from the mains. In contrast, the PCIe-6255 is mains-referenced, as you would have installed it in a desktop powered by AC mains.

 

The noise spikes you see are not inherent to the DAQ but rather the external connections, easy way to check by shorting the AI+ and AI- to AIGND and capture the signal to understand the inherent noise of the DAQ.

 

Please read the following,

https://www.ni.com/en-us/support/documentation/supplemental/06/field-wiring-and-noise-considerations...

 

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos