Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

Voltage drift while using ADC in USB 6341

Hello, I am using NI DAQ USB 6341 in ADC (channel AI0) for recording analog signal from Thorlabs PDA36A2 power diode. This PD has a BNC output and i tried connecting the negative terminal to ai0- and aignd. However, when i use RSE mode, the voltage keeps on increasing and then saturates. In differential mode, the voltage increases and then decreases again so after a minute, the data seems to be ok but it is not absolute value, which is what i require. Cant seem to fix the issue by changing connection of the negative pin from ai_gnd to ai_0-.  

MountainEater_0-1745591421522.png

 

0 Kudos
Message 1 of 4
(548 Views)

Maybe you need some bias resistors?

0 Kudos
Message 2 of 4
(528 Views)

Probably wants a 50 ohm load instead of the high input impedance of the DAQ.

Try putting a 50 ohm resistor across the two input terminals of the DAQ.

 

-AK2DM

~~~~~~~~~~~~~~~~~~~~~~~~~~
"It’s the questions that drive us.”
~~~~~~~~~~~~~~~~~~~~~~~~~~
0 Kudos
Message 3 of 4
(516 Views)

@AnalogKid2DigitalMan wrote:

Probably wants a 50 ohm load instead of the high input impedance of the DAQ.

Try putting a 50 ohm resistor across the two input terminals of the DAQ.

 

-AK2DM


You're spot on with the guess. The user manual states that you need a load resistor and selection of the load also impacts the response.

santo_13_1-1749717684971.png

 

Are you using USB 6341-BNC or the breakout version?

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 4 of 4
(312 Views)