Weird Analog Input voltage reading
Hi Guys,
This is a really weird problem I am not able to figure out. And it is with respect to Analog Input voltage reading into DAQ 6008. I have a voltage reading of 1.7into ai5 and GND (RSE). Reading the voltage directly from the sensor (the device that gives this 1.7V) I get a reading of 1.69V. When I connect it to the DAQ and read the voltage off the screw terminals (Using DMM) , I read a voltage of 1.55V, and offset of 0.14V.
Now, when I disconnect the DAQ USB cable from the DAQ board and then measure off the screw terminals, I get the right 1.69V. But, when I put the USB back on the DAQ card, I get the 0.15V offset again. What is going on here?
V
I may not be perfect, but I'm all I got!
(3,568 Views)
Try reading it differential to see if the problem goes away. You may have a grounding issue.
Richard
(3,560 Views)
What is the sensor exactly? Do you know how much current it can source? The reason I ask is because I think the analog inputs on the 6009 have an input impedance of only 140kohms.
-AK2DM
(3,553 Views)
analogKid is dead on!  so 0.15V/1.7V so you are loading your signal by about 9%  I guess the source impedaince would have to be about 15Kohms
Message Edited by Jeff Bohrer on 03-19-2010 05:27 PM
(3,530 Views)