What is the real current rating of the PCI-6259 dio?
The spec of the PCI-6269 daq board suggests that the DIO should be able to generate around 20mA of current, but I am finding that if I feed it into a single DIO into a 1kohm lowpass RC filter (sample rate of 1MHz), the filter seems to be significantly loading down the signal.  This should only draw about 5mA.
Is the 20+mA a safety spec?  If so, what is the actual current source/sink capability of the DIO?
Sean
(3,109 Views)
Hi Sean,
The maximum current you can source/sink with your PCI-6259 is 24 mA. Can you please explain what you mean when you say that the filter seems to be significantly loading down the signal?  Are you seeing less than 5 Volts for DO high level? Are you using multiple lines in the port to source current to multiple external circuits?
Wallace F.
(3,093 Views)
Posting to be informative and give fair warning:
I meant to write up something more comprehensive, but life is busy...  😉
I am using the dio of an m-series daq card to generate additional "analog outputs" by generating delta-sigma bitstreams (followed by hardware filters and amplifiers).  I was having trouble because it seemed as though the signals were getting loaded down.  After investigating what was going on:
1) The daq dio was performing as specified.
2) The output of the filters were being loaded down by the inputs of the op-amps being used.
It turns out that the op-amps have diode-tied inputs (presumably as a protective measure?).  These were not described anywhere in the text of the spec sheet, but I found them in the transistor diagram.  In many op-amp circuits this would not be a problem, because the amplifier gain is used to drive the difference between the inputs to very small levels, but in my case, I was using them as integrators in a feedback loop (with the signal from the PC as a PC-controlled reference/set point value.
Since this seems to be true of more than a few op-amps, (I checked several other datasheets), I thought it might be useful to give warning to others who may be encountering similar problems.
Cheers!
(3,021 Views)