Where do I find data on the gain accuracy and gain stability of a NI6225?
I need to answer the following two requirements;
Gain Accuracy: The gain accuracy shall be at least ±1% of the selected dynamic range.
Gain Stability: Thirty minutes after applying power to the system, the gain stability shall be better than 0.33% per minute. The total change in 1 hour shall not exceed ±3%.
(3,557 Views)
Accepted by topic author etechstan
Hi etechstan
Heres the link for the technical specification sheet for the NI 6225:
http://sine.ni.com/ds/app/doc/p/id/ds-15/lang/en
The gain accuracy depends on your selected range and using the formula from the table below you can calculate the Absolute Accuracy and Gain and Offset Error.
AI Absolute Accuracy Table
Nominal Range Residual
Gain
Error
(ppm of
Reading)
Gain
Tempco
(ppm/°C)
Reference
Tempco
Residual
Offset
Error
(ppm of
Range)
Offset
Tempco
(ppm of
Range/°C)
INL Error
(ppm of
Range)
Random
Noise,
σ (µVrms)
Absolute
Accuracy
at
Full
Scale1 (µV)
Sensitivity2
(µV)
Positive
Full
Scale
Negative
Full
Scale
10 -10 75 25 5 20 57 76 244 3,100 97.6
5 -5 85 25 5 20 60 76 122 1,620 48.8
1 -1 95 25 5 25 79 76 30 360 12.0
0.2 -0.2 135 25 5 80 175 76 13 112 5.2
AbsoluteAccuracy = Reading · (GainError) + Range · (OffsetError) + NoiseUncertainty
GainError = ResidualAIGainError + GainTempco · (TempChangeFromLastInternalCal) + ReferenceTempco · (TempChangeFromLastExternalCal)
OffsetError = ResidualAIOffsetError + OffsetTempco · (TempChangeFromLastInternalCal) + INL_Error
1 Absolute accuracy at full scale on the analog input channels is determined using the following assumptions:
TempChangeFromLastExternalCal = 10 °C
TempChangeFromLastInternalCal = 1 °C
number_of_readings = 100
CoverageFactor = 3 σ
For example, on the 10 V range, the absolute accuracy at full scale is as follows:
GainError = 75 ppm + 25 ppm · 1 + 5 ppm · 10     GainError = 150 ppm
OffsetError = 20 ppm + 57 ppm · 1 + 76 ppm       OffsetError = 153 ppm
AbsoluteAccuracy = 10 V · (GainError) + 10 V · (OffsetError) + NoiseUncertainty AbsoluteAccuracy = 3,100 μV
2 Sensitivity is the smallest voltage change that can be detected. It is a function of noise.
Accuracies listed are valid for up to one year from the device external calibration.
If you have any furthur questions please do not hesitate to contact me.
Kind regards
David
National Instruments UK
Many thanks for using the NI forums and I look forward to your reply.
Kind regards
David
(3,542 Views)
