03-31-2008 08:47 AM
I am measuring a 1 kHz stepwise signal at a sampling frequency of 200 kHz using a PXI 4461 board.
The result is not stepwise at all and there are oscillations before and after each step oscillation (see jpg-file)!
If I can understand oscillations after the step transition (cable reflection or other) I can not figure out why there are also oscillations before the step transition!
Moreover, these oscillation are not present on the oscilloscope!
Is it an artifact of the sigma-delta conversion algorithm or a defect of my board?
Does anybody explain to my the origin of these oscillations?
I thank you in advance
Frédéric

Solved! Go to Solution.
04-01-2008 03:11 AM
04-01-2008 06:52 AM
Hi Laurent,
Thanks for your answer but it does not fix my problem. I do understand that one can not generate a perfect stepwise signal with a small (200 kHz) bandwidth. However, in my case, the signal is not generated by the 4461 board but by an external source (an ac Josephson voltage array) and the signal is indeed almost stepwise (with few tens of nanoseconds of rise time). On the other hand, even with a cheap oscilloscope, I would be able to see the 50 mV oscillation! Therefore, the oscillation is definitively not included in the generated stepwise signal.
What I am doing with the board is only to use the ADC input and I am still looking for the reason of such oscillation on the sampled results…
Do you have any other suggestion?
Best regards,
Frédéric
04-02-2008 09:47 AM
04-02-2008 10:17 AM
Hi Frédéric,
Do you know how your „ac Josephson voltage array“ external source generates square waveforms. If they are done out of sinus as I suppose, the antialiasing filter present in the PXI 4461 can remove some harmonics.
Explanations:
Indeed, according to the specifications, the sampling frequency (Fs) of
the PXI 4461 is 204,8kHz. Between the input and the DAC, there is an
antialiasing filter (low pass filter) which cuts all the frequency above Fs/2;
So in your case, 100kHz. As a result just the harmonics under 100kHz are taken
into account and that’s for sure the reason why the square has these
oscillations.
To check this theory, try first with a higher frequency you should have more oscillations, and then try with lower frequency you should have less oscillations.
The documentation of the PXI 4461 board can be found here:
24-Bit, 204.8
kS/s Dynamic Signal Acquisition and Generation
http://www.ni.com/pdf/products/us/pxi4461.pdf
Best regards,
Laurent
04-02-2008 10:19 AM
04-02-2008 10:25 AM
04-02-2008 10:47 AM
@Ov_metas wrote:Therefore, for a 50 Hz signal I would have almost 2000 harmonics and the measured signal should be steeper. Is it right?
04-02-2008 01:20 PM
04-07-2008 12:53 PM
Does the delay Chris is speaking about the 63 samples delay?
Can we turn off this digital filter?