Hello Hgmittal,
The NI-9237 module uses a sigma-delta analog to digital converter to perform its measurements. As a result, it requires an oversample clock and a derived sample clock to operate. Check out this forum post or more information. This post suggests using an NI 9219 for the strain task since it can accept an external sample clock for a timing source. If you have an external excitation source, you can use your 9215 to perform a strain measurement with an external sample clock as well.