Hello,
What exactly do you mean by �connect in series?� If you're looking to cascade the RTD's, this is not supported by the SCC RTD01.
I do have a couple of recommendations, however, for improving the accuracy of your measurements. First, you can over-sample the incoming signal and average the data. Second, you can upgrade to one of our SCXI devices. The SCXI module will provide higher accuracy than the SCC module, and you will want to use a SCXI-1102 coupled with a SCXI-1531 for current excitation.
I hope this helps! Please let me know if you need further assistance. Have a great day!
Kind Regards,
Joe Des Rosier
Nation
al Instruments
03-10-2004 11:06 AM
03-10-2004
03:49 PM
- last edited on
11-14-2025
12:17 PM
by
Content Cleaner
Tom,
After reading your second post, I realize that you want to couple the RTD's in series. Originally, I thought you wanted to couple the modules in series.
Ok, you can place the RTD's in series, but this will not increase your measurement accuracy. If you decide to experiment with this setup, make sure you provide sufficient current excitation for both RTD's. What will happen is the experimental error you receive at the first thermocouple will be seen by the next RTD in series, so your error will not decrease.
I advise you to try the oversampling idea to increase your accuracy.
You can use this accuracy calculator to calculate the accuracy of a given SCXI module and compare this value to the measurement accuracy you're seeing with the SCC module.
I hope this helps! Please let me know if the averaging technique helps to increase your accuracy.
Best of Luck,
Joe Des Rosier
National Instruments