analog input DAQ6020E (BNC)
hi,
I am trying to measure a voltage signal that is between 30-40V with a DAQ 6020 E(BNC) device. What is the maximum voltage that i can safely measure at the analog input channels? I think table 4-1 on the DAQ 6020  E (BNC) manual says the voltage limit is 35V on and 25V off. Is that correct?  What if my signal is bigger than 35 V?
(3,312 Views)
You need to read the specs a bit more carefully. They clearly state that the +/-35 and +/25 are the overvoltage protection. The input signal range is +/-10V.
(3,304 Views)
poli,
If you are worried about your voltage level then you can use our Signal Conditioning hardware to address this concern.
NI SCXI-1104 : This module is part of our SCXI series.  It will require a chassis and connector block.
NI 9229 : This module is part of the CompactDAQ line, it will come with a USB connection sleeve.
I hope this helps!
Ryan N
Application Engineer
ni.com/support
(3,276 Views)