cDAQ-9137 RS232 RJ45 Port
I try to use the RS232 RJ45 port on cDAQ-9137 chassis to comunicate with CSZ thermal chamber.
I use a RJ45 to DB9 converter (Digikey Part # 602-1506-ND) with a null-modem (crossover) cable (Digikey Part # TL642-ND) to connect between cDAQ and CSZ thermal chamber.
The CSZ thermal chamber use Modbus and only pin 2, 3, and 5 is present. I made the labview program to talk to the device and tested using my laptop with USB-COM converter. and the program works with no error.
Once I switch to cDAQ and use the same program. The error in MODBUS read presents. I shut down both cDaq and thermal chamber and restart them but it still won't work.
I used COM1 in cDAQ for this communication testing. The cDAQ also used a Virtual COM3 (VCP) to communicate with other device.
Seems like the port on cDAQ has some problem or I didn't configured it properly.
Any recommendation and advice is appreciated. Let me know if further information is required.
(4,728 Views)
Hi Luyi,
Have you made sure to have installed NI VISA in your cDAQ?
Daniel C.
(4,681 Views)
Hi Daniel,
Thank you for the reply. in NI MAX, the COM1 is visible.
Eric
(4,674 Views)
Hi Eric,
Try making a loopback test on the cDAQ's serial port to make sure it is configured properly.
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000YFtECAW&l=en-US
Daniel C.
(4,659 Views)
Thank you for reply.
I compared the pinout from cDaq manual with the converter cable from Digikey. They are not the same.
cDaq uses 8 and 9 for TxD and RxD. However, EIA-232D dictates 5 and 6 for TxD and RxD.
I'm ordering converter cable from NI (#182845-01) to fix this. This is not the first time that a simple regular hardware off the shelf somewhere won't work on NI.
I will get the cable in a week and will try it out. Thank you all for the help.
(4,656 Views)