There are a number of troubleshooting measures to consider at this point.
When the 6015 driver (NI-DAQmx) is installed it also installs an application called Measurement and Automation Explorer. This utility is desiged to help users test their hardware and the I/O on their device. Please make sure you have the latest version of DAQmx (8.5) installed and open this utility. Find your card listed under Devices and Interfaces, right click on it and select 'test panels.'
When you are connecting a signal to Ai0 with the test panel running in differential mode, please make sure that you are using the pins for ai0 and ai8. You can actually create an analog acquisition task in MAX and click the Device Routes tab at the bottom to see the necessary connections. This will probably be pins 34 and 68. Confirm that you've set the card up this way, run the test panel and let me know if you're still having problems.
If you still experience the same issue, could you reply with the actual values you're reading versus the values you're expecting? What do you mean by 'more correct?'
Elijah Kerry
NI Director, Software Community