ni 9219 _ measurement uncertainty _ sensitivity
I'm using a NI 9219 in current mode to acquire several signals, and I need to calculate the measurement uncertainty due to the module.
In order to do so, I must know the resolution, the absolute accuracy and the sensitivity.
According to the manual (specifications start page 23) :
- Resolution is 24 bits
- Absolute accuracy is function of gain error and gain offset (For typical use : Absolute Accuracy = Gain error * measured value + gain offset * full range)
- Sensitivity is ???
How can I know the sensitivity of the module in current mode ?
Thank you & Regards
Noé
(4,151 Views)
Hi Noé,
If the sensitivity is not specified then it should be calculated dividing the range you use by the number of possible values.
As you use a 24 bits the result should be something like : (25 - (-25)) mA / 2^24 values.
I hope it helps even though I may be mistaking.
Maxime L
Certified LabVIEW Architect
(4,085 Views)