Hi Gilles,
I tried to find out what could be wrong with your system. I found the following knowledgebase entire, which may assist you in setting up your Application. I have not read in your Post, that you have tried several ranges in your application. I would try this first and set the Input range as small as possible to the Temperature you expect.
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P8wxSAC&l=en-US
Hope this helps, if not don´t hesitate to answer to this Forum-Entry again.
Alex
NI Switzerland
03-21-2006 07:57 AM
04-25-2008 02:37 AM
04-28-2008
11:10 AM
- last edited on
03-14-2025
04:49 PM
by
Content Cleaner
Hi,
The sensitivity of the K type thermocouple is about 41 µV/°C. For the 6052E, at the minimum range, the sensitivity is 35uV, which should still not be causing too much noise but it is close. External noise and the accuracy of the card might cause variations. Could you try reading in a voltage measurement only from the thermocouple at the smallest range and see what kind of signal you receive? Have you considered some kind of signal conditioning system, or maybe just an amplifier. For example the 1102 scxi is designed for thermocouple measurements.
Regards,