pci 6221 analog out voltage flat tops above 1V
I have a pci 6221 DAQ card going into a 2090 breakout box and I am trying to get a 20 volt peak-to-peak out of the analog out.  Using the MAX test program I can only get 1 V out of analog output 1, anything higher than this and the signal flat tops, i.e. I don't get the "peak" of the sine wave when looking at in on an oscilloscope.  When I run it in the -10 to 10 V range the signal looks more like a square wave than a sine wave.  Analog out 0 is even worse and I can only get 0.5 V before the same thing occurs.  I know it's not the scope because I can feed in the output from a waveform generator and get the -10 to 10V sine wave.  In a vi I can only get a 200 mv sine wave, anything more and I get an error that the max output voltage has been exceeded.  I tried to reset the max voltage in the vi but get an error that I can't edit an existing task.
Is this indicative of a software problem or a hardware problem?
 Sorry if this doesn't make too much sense, I'm pretty new to labview DAQs.
(3,315 Views)
Is your scope set to 50Ω input impedance?  If so, the 6221 outputs are going into current limit.  You should be able to switch your scope to 1MΩ input impedance.
(3,310 Views)
switching to 1M input impedance fixed it Thanks!
(3,303 Views)