Hi Rodrigo Schneiater.

 

 

Thank you for answer me. I'll clarify the scenario for you.

 

Beyond the hardware that I had listed before, we have a scxi 1305 terminal block for current signal input.

 

We did some tests, in a frequency range between 60 up to 400Hz, like you sugested me. We used a current clamp, that have a constant phase error for all range of the test (comproved by oscilloscope measurements). This is the result for, discounted the error by the current clamp (3°):

 

 

60Hz - phase shift of 6.40° between voltage and current signal;

 

120Hz - phase shift of 11.15°;

 

180Hz - phase shift of 16.50°;

 

240Hz - phase shift of 21.66°;

 

300Hz - phase shift of 26.40°;

 

360Hz - phase shift of 30.70°;

 

400Hz - phase shift of 33.30°;

 

 

 

Thank you for help us.