Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

problem of analog input with PCI-6115 and BNC-2110

Solved!
Go to solution

Hi, I have a DAQ PCI-6115 card and BCN-2110 connector that I want to continuously measure the analog voltage. Now I can continuously measure the laser light intensity by photodiodes through channel 1, and get the reference through channel 0. Then I can get the harmonic signals by the MultiChannel LockIn - DAQmx.vi. However, I also want to simultaneously detect another analog signal in voltage throuth channel 2. The problem is: the voltage range of channel 2 is about 0.3 mV to 6 mV which is far lower than the other 2 channel (0-10V), how could I set different limitation per channel to continuously and simultaneously acquire all data?

 

I found Dennis Knutson provided the solution a year ago, that is shown below:

 

Multiple Channel Config.PNG

 

In the solution, one would have arrays of channels, ranges, configurations, but I don't know how to apply it to my case. Is there any suggestion or other solution?

 

Thanks a lot!

0 Kudos
Message 1 of 8
(5,174 Views)
歡迎致電 02-2377-2222 工程部
0 Kudos
Message 2 of 8
(5,148 Views)

Dear Charles0820, thanks for your suggestion, I will take your advice. 

 

But I should detailed my problem first. The vi is attached, and some part of it is shown below:

 

fig.JPG  

 

 

However, it seems to be wrong cuz the message shows the specified resource is reserved. Please gime me any suggestion, and I do appreciate your help!

0 Kudos
Message 3 of 8
(5,144 Views)

Why didn't you try to search for the error code or error message. You would have gotten numerous posts on this basic mistake.

 

The answer is simple - you simply cannot run separate tasks on the same hardware resource. Use a single DAQmx Read and you will get both channels.

 

And when you rewrite the program, remove the stop and clear tasks from inside the loop.

0 Kudos
Message 4 of 8
(5,138 Views)

I revised the vi as following:

 

fig1.JPG

 

But only the last task is output from the for loop, so I can merely acquire the voltage from physical channel 2. I tried to enable the indexing of the output task, so both task can be output, but it failed to input more than one task into the DAQmx timing.vi.  Could you please give me more suggestion? What I need is to continuously acquire the voltage from channel 0 and 1 to decompose the signal, and also acquire the voltage from channel 2. The problem is the voltage from channel 0 and 1 is much larger than that from channel 2, so if I set the voltage range of both task in the same scale, I can't resolve all signals.

 

Thanks a lot!

0 Kudos
Message 5 of 8
(5,114 Views)
Solution
Accepted by topic author Ivan777

You've got the task output autoindexed and that is not correct. Try the code below.

 

Message Edited by Dennis Knutson on 01-21-2010 07:59 AM
Message 6 of 8
(5,108 Views)

It works!! Thank you very much, Dennis. You truely do me a big favor and help me to understand DAQmx further. May I ask one more question? Why can't I use the syntax "Dev1/ai0:1"? I thought I can use this syntax cuz the voltage range of channel 0 and 1 is the same.

 

fig2.JPG

0 Kudos
Message 7 of 8
(5,091 Views)

Sorry, the reason should be the auto indexing.

 

 

 

 Thank you again!

0 Kudos
Message 8 of 8
(5,088 Views)