range 9219
    What is the best (most accurate) measurement range of 9219. I plan to use a thermistor with 9219 to measure temperature. The thermistors has different resistance selections. I want to choose the type which matches the 9219. At what resistance range the 9219 read the most accurate reading (smallest percentage of error)?
ryan
(4,570 Views)
Hi Ryan,
I think what we need to consider here is not so much the resistance of the thermistor, but rather how that resistance will affect the voltage range. The 9219 has different voltage ranges from 125mV to 60V. You will need to choose a current and a resistance that takes full advantage of one of those ranges to get the best meausurement. So, just applying the old V=IR should do it!
Hope this helps!
Margaret Barrett
Digital Multimeters and LCR Meters
(4,543 Views)
Ryan,
You may also want to review the NI 9219 module specificaions here: 9219 Specifications
10k and 1k 4- and 2- wire resistance gain error is 0.1% typical/0.5% max.
(4,504 Views)
Thanks. Your field engineer loaned me one and I am testing it right now.
Ryan
(4,493 Views)