removing usb-6009 floating voltage
hello,
I have created a VI that logs 6 voltages from 6 seperate batteries, and all is well, until i disconnected one of the batteries and it stills sees approx 1.4 volts. how can i remove this offset(?)
Is it present on the connected readings?
Regards
Steven
(4,246 Views)
You are probably seeing what is called 'ghosting'. There are numerous threads on this. You can do a search for this. One solution is to not leave any inputs floating.
(4,243 Views)
i have to test batteries down to below 1.0 volts, will this ghosting be present then with a 1ma load?
Steven
(4,239 Views)
Accepted by topic author mgbglasgow
The USB-6009 has a non-standard input circuit which is terminated into ~1.4 V through ~144000 ohms. Check the manual for a schematic.
As long as your source has an impedance much lower than 144000 ohms and can sink current (if the source voltage < 1.4V), then you should be OK.  Batteries, even nearly discharged, should not be a problem.
Lynn
(4,229 Views)
