ripple/noise and faulty corrections acquiring with PXIE-4496 from impact hammer
First: What do you want to measure?
What are the specs of your hammer (incl. sensor)?
Usually you measure something on the target (anvil 😉 ) too.
The FRF of the 4496 IEPE input for the low frequency is coarsly a 33nF, 10M high pass that coudn't be turned off and should take about 5.5s to settle (it's all in the manual 😉 )
Greetings from Germany
Henrik
LV since v3.1
Message 11 of 17
(1,722 Views)
I am doing dynamic identifications of beam elements, so on the beam I have several IEPE accelerometers with which I measure  the response to the impact. With impact and accelerations, I build FRFs for different points.
The hammer is a Bruel and Kjaer 8206-001 IEPE impact hammer. 11.4 mV/N, <100 ohms output impedance,
thanks for the 5.5s tip!!
Message 12 of 17
(1,719 Views)
1. What does the response of a beam accelerometer very close to the impact point look like? Preferably a massive, stiff beam.
2. What does the hammer response look like if you turn it upside down and hit the underside of the beam, striking upward?
Lynn
Message 13 of 17
(1,707 Views)
The response of a concrete stiff beam looks, I think, nice (picture below + detail yet below), gets damped and goes back to zero. As for the hammer upside down, I haven't done it. I will do it next week and post. Both hammer and accelerometers are IEPE.
Message 14 of 17
(1,701 Views)
This makes me think that there is nothing fundamentally wrong with your setup.  Something either mechanically or electrically is slowly resetting to the queiscent state.  It just takes much longer than you are expecting.
Lynn
Message 15 of 17
(1,693 Views)
Just as promised, here is the response of the hammer upside down. Sampling frequency is also different (down from 50000 to 25000 Hz). Again, the oscillations are only a 0.2% of the impact magnitude, so I guess no real problem. Thanks.
Message 16 of 17
(1,674 Views)
Accepted by LAPadron
As you say, the signals which you cannot explain are very small.  Unless it becomes a real problem for your analysis, try not to worry about it too much.
Lynn
Message 17 of 17
(1,669 Views)