synchronizing analog output and digital output
I just want to output waves from the USB6421 ao0 and DIO0, but I find there is a delay between the analog and digital waves. I generated the  analog and digital waves with same sampling rate.
Generated analog and digital waves:
Output analog and digital waves by USB6421 measure by Oscilloscope:
This my code
import nidaqmx
import numpy as np
from nidaqmx.constants import AcquisitionType,Edge
import matplotlib.pyplot as plt
# 参数配置
fs = 1000
duration = 1
num_samples = fs * duration
wave_amplitude = 2.0
wave_period = 0.5
symmetry = 0.5
t = np.linspace(0, duration, num_samples, endpoint=False)
# Generate the Triangular Analog wave
phase = (t % wave_period) / wave_period
rise_samples = int(wave_period * fs * symmetry)
fall_samples = int(wave_period * fs * (1 - symmetry))
galvo_data = np.zeros_like(t)
for i in range(len(t)):
    cycle_pos = int((phase[i] * (rise_samples + fall_samples)) % (rise_samples + fall_samples))
    if cycle_pos < rise_samples:
        #from -A to +A
        galvo_data[i] = wave_amplitude * (-1 + 2 * (cycle_pos / rise_samples))
    else:
        # from +A to -A
        galvo_data[i] = wave_amplitude * (1 - 2 * ((cycle_pos - rise_samples) / fall_samples))
# Generate Digital Square Wave
dove_data = (t % 0.5) < 0.25
plt.figure(figsize=(10, 4))
plt.plot(galvo_data)
plt.plot(camera_data)
plt.show()
with nidaqmx.Task() as ao_task, nidaqmx.Task() as do_task:
    ao_task.ao_channels.add_ao_voltage_chan("Dev1/ao0")
    ao_task.timing.cfg_samp_clk_timing(
        rate=fs,
         source="",
         active_edge=Edge.RISING,
         sample_mode=AcquisitionType.CONTINUOUS,
         samps_per_chan=num_samples
     do_task.do_channels.add_do_chan("Dev1/port0/line0")
     do_task.timing.cfg_samp_clk_timing(
         rate=fs,
         source="/Dev1/ao/SampleClock",
    ao_task.write(galvo_data)
    do_task.write(camera_data.astype(np.bool))
    ao_task.start()
    do_task.start()
    input("press stop")
    ao_task.stop()
    do_task.stop()
(676 Views)
What's missing is that you need to sync the tasks to start at the same time.  Right now there's an offset due to the execution time needed to get the tasks started under software timing control.  You need to sync their start times via hardware instead.
Because both your tasks are on the same device, a "start trigger" will be sufficient.  My recommendation: configure both tasks for a start trigger from one of the PFI terminals that serves double duty as a DIO line.  Then even after the software calls to start the tasks, signals won't start generating your waveforms until the hardware trigger signal asserts.
You can control the trigger with a separate DO task that you'll start *after* starting the other two tasks.  A quick toggle high and low will make your AO and your other DO get started in sync.  The fact that they're on the same device means that they derive their sample clock from the same internal clock so they'll *stay* in sync.  (This wouldn't be true if they were on different devices.  Small tolerances in clock frequency would make them drift apart over time.  A pretty common clock spec would allow an error of about 3 millisec per minute of acquisition.)
-Kevin P
(649 Views)
Some indication to port into python the solution explained in previous post:
I note that you are configuring "source" through cfg_samp_clk_timing. If you had 2 or more separate devices and a connection between the 2 through RTSI or PFI, the framework would share the clock from master device to slave device and they would stay in sync (I am doing clock sharing through RTSI). Here you are on same device so works without extra cable
To share the start trigger, I am doing:
for slave_task in all_slave_tasks:
   slave_task.triggers.start_trigger.cfg_dig_edge_start_trig("/" + master_task.devices[0].name + "/ai/StartTrigger") # this will be /Dev1/ai/StartTrigger, you need to adapt to your terminal, like ao or do. Well, no do/StartTrigger in https://www.ni.com/docs/en-US/bundle/ni-daqmx/page/termnames.html so ao may need to be the trigger or here is some concept I don't know
Then:
   slave_task.start()
master_task.start()
slave_task will start upon receiving master_task trigger
(635 Views)
Thanks for interpreting into Python, ft_06!
To the OP: the method shown by ft_06 that triggers the slave task from an internally-generated signal of the master task is likely the method I would use myself.  I described the completely separate external trigger method only due to guessing that it might be more easily understood and translated into Python.  But since ft_06 just did the important parts of that translation, go ahead with his/her method.
-Kevin P
(618 Views)
thanks for your replay, I solved the problems
(591 Views)
