thermistor MAX
I' m trying to calibrate and take measurements with a thermistor YSI 44033. I' m using PCI6034E, connector block SCXI-1121 and terminal block SCXI-1321. I' m creating in MAX a NI-DAQmx with internal current excitation and 2-wire configuration, when I' m trying to run the project, I have an error -200077 occurred at DAQ Assistant. I have set excitation of scxi1121 and scxi1321 in current mode by replacing the jumpers at the board as I have read in the manual.
error -200077:
Error -200077 occurred at DAQ Assistant
Possible Reason(s):
Measurements: Requested value is not a supported value for this property.
Property: AI.Max
You Have Requested: 100.000000
Valid Values Begin with: 160.641471
Valid Values End with: 245.871833
Channel Name: Temperature
I would be grateful if you help me!
(4,424 Views)
The property AI.Max is in scaled units (in this case Degrees C) Write a value greater than 160.xxx
(4,409 Views)
Thanx Jeff,
The problem is that even if I write a value greater than 160,xxx and lower than 245,xxx, the result of the measurement it's always -273,15!
I think that the problem may be in the thermistor coefficients or in the calibration of the SCXI1121. I' m right? Any other idea?
Thank you for you help
(4,398 Views)
-273.15 C = 0 Kelvins.  So you are reading a properly scaled 0V input and the calibration seems pretty close.  Some physical wiring is not correctSmiley Sad but the BD Wiring seems to work.Smiley Happy
(4,395 Views)
Hi schi86,
Check the SCXI-1121's gain jumpers and the corresponding module properties in MAX. If the gain is too high, that limits the voltage range and consequently the range of temperatures you can measure.
Brad