usb-6009 low resolution(for mV range) problem
Hi
I am using NI USB-6009 for a temperature measurement application. My analogue voltage input range is 0-100mV with 0.1667mV/°C. But the resolution of my measured values is  1-2mV. I changed the input range form -5 to5 Volt to 0 to 150 mV but I didn't see any change. Also averaging is not helping me a lot since I want an accuracy of less than 0.1667mV. I am also using differential mode. I read that this devices can not be calibrated.... What are the options to get an accurate result.
Tnx
(3,643 Views)
Have you looked at the specs?
You options would seem to accept the measurements you are getting or use a board that has much greater accuracy.
(3,642 Views)
Cherkose,
U should be able to set the differential range to -1 to 1 V, I don't think setting your range lower than this will make a difference, your range will probably just coerce to -1 to 1V... 14 bits should give you a 0.122mV resolution on the ADC then.
(3,625 Views)
While 14-bits across the ±1V range will give a code width of 0.122 mV, code width is only one factor in determining the accuracy of a measurement.  Many other factors, such as noise & temperature add to inaccuracy.  As such, it's always best to refer to the absolute accuracy specifications listed in the Detailed Specifications for your device.  For the ±1V range of the USB-6009, the absolute accuracy is ±1.53 mV at 25°C.  As you move away from 25°C, this will get worse.
The USB-6009 is not designed to measure thermocouples, as thermocouples require high accuracy and precision.
Seth B.
Principal Test Engineer | National Instruments
Certified LabVIEW Architect
Certified TestStand Architect
(3,614 Views)