06-04-2013 05:46 AM - 已編輯 06-04-2013 05:51 AM
您好.
我使用pxi-1033機箱,solt 2插 pxi-4132
slot 4插pxi-4071
跑CALIBRATION EXCUTIVE 3.5 執行verify only ,
在current measure時(4132 HI 接4071 從上數下來第3孔, 4132 LO接4071 從上數下來第2孔)
出現下列錯誤訊息:
Error -1074118642 occurred at IviDmm IVI Error Converter.vi
Possible reason(s):
The resolution you have requested is invalid for the current range.
Parameter 4 out of range, or error occurred while setting Parameter 4.
Resolution
Complete call chain:
IviDmm Configure Measurement.vi
cex_Instr_IVI_DMM_Configure.vi
cex_Instr_DMM_Configure.vi
Verify_NI 4132_Current Output Dmm Communication.vi
Verify_NI 4132_Current Output.vi
Verify_NI 4132_Current Output.vi.ProxyCaller
Error -2147352571 occurred at 類型不相符。
in cex_ConvertUserField.vi->cex_ConvertUserFields.vi->cex_ConvertCalData30ToCalData.vi->cex_ConvertCalData30ToCalData.vi
at step Convert CalData30 to CalData.ProxyCaller
This error code is undefined. Undefined errors might occur for a number of reasons. For example, no one has provided a description for the code, or you might have wired a number that is not an error code to the error code input.
Additionally, undefined error codes might occur because the error relates to a third-party object, such as the operating system or ActiveX. For these third-party errors, you might be able to obtain a description of the error by searching the Web for the error code (-2147352571) or for its hexadecimal representation (0x80020005).
請幫忙解決, 是否不能用4071. (7.5 digits)
NI PXI-4132 Calibration Procedure上面寫
DMM :Agilent 3458A (or using at least 7.5 digits)
4071應該解析度是ok.
謝謝.
已解決! 轉到解決方案。
06-04-2013 06:01 AM - 已編輯 06-04-2013 06:02 AM
http://www.ni.com/pdf/manuals/375129a.pdf
4130 calibration procedure 推薦 4071
4132 並沒有
at least 7.5, 有可能需要到 8.5, 所以沒有推薦 4071
在 06-04-2013 08:31 AM
Hi Peter,
This error is being generated because Cal Exec is trying to set your DMM's resolution to 7.5 digits for a DC current measurement. When taking DC current measurements, the resolution of the PXI-4071 can only be set to a maximum value of 6.5 digits.
Do you have an Agilent 3458A (or equivalent) available to use for the verification process?
06-04-2013 07:55 PM - 已編輯 06-04-2013 07:56 PM
Thanks a lot.
I think I should buy an Agilent 3458A.