Nick,
Thanks for the reply. This problem went away when I opened the IVI example sequence that is installed along with TestStand. It has not repeated since. I can now access my defined 'TekScope'.
I am having a problem getting the IVI code to run in simulation mode. I modified the TestStand 3.1\Examples\IVI\IviStepTypes.seq file to change the Fileglobals.ScopeLogicalName value to TekScope which is the scope name in the dropdown box for edit IVI tools. I then selected the scope steps and selected Run Selected Steps. I get the following error:
An exception occurred calling 'RunStep' in 'ISubstep' of 'TestStand Ivi Step Types'
An error occurred while executing the step.
IVI Error: The IVI instrument failed to initialize for logical name 'TekScope'.
Details: TekScope_InitWithOptions returned an error: -1074134964 (0xbffa004c). The option string parameter contains an entry with an unknown option value.
Source: TSIviStepTypes
-1074134964; User-defined error code.
In MAX the TekScope.DriverSession - Simulate With setting is nisScope, Driver Setup is blank.
01-28-2008 02:02 PM