More information about the problem. The error dialogue reads as follows :-
error -200546 occurred at Daqmx Close External Calibration.VI at step adjust.
Possible reason(s):
Measurements: External calibration failed,and the external calibration data has not been changed.
Make sure that the reference signal used for the calibration is stable and that the voltage matches the value specified to the calibration software. Disconnect the device from any external signals that might be introducing noise.
Physical channel name: SC1Mod1/ai26.
As far as I can tell from my instruments the voltage supply is stable. I seem to have problems evey time I start to use the harware/software (usually once or twice per year). Am I doing anything wrong?
cheers
Herbie
03-03-2008 04:02 AM
03-03-2008 04:07 AM
Hi
I am using Daqmx 8.5: traditional Daq 7.4.1 is also installed on the PC.
cheers
Herbie
03-03-2008 04:32 AM
03-03-2008 04:38 AM
Hi
I will try your suggestion and let you know how I get on.
Cheers
Herbie
03-04-2008 05:18 AM
Hi Dan
Installing Daqmx 8.7.1 has helped with the error situation.Thanks for that. My SCXI1102C modules are still failing to calibrate. I have noticed that the value shown on the DMM (coming from the module) is different to that shown on the screen by at least 1-2 mV. This causes values deemed to be within the required limits by the DMM are classed as not meeting the limits by the software. Any thoughts on this matter?
cheers
Herbie