PXI
5922 5124 Fuse rating , output spec of 5V @ 9-pole mini DIN AUX connector
cross posting from , this board seems to more active ...
I can't find the specifications of the +5V output current rating (Pin 1 to Pin 2 GND) on the scope cards (5922/5124)  9-pole mini DIN connector
Greetings from Germany
Henrik
LV since v3.1
(5,999 Views)
Hi Henrik,
just to clarify your question: You want to know the maximum output current for the +5V Pin, not for the PFI0/PFI1 pins?
NI PCI-5124 Front Panel - NI High-Speed Digitizers Help - National Instruments
https://www.ni.com/docs/en-US/bundle/legacy-scope-hardware-features/page/ni-pci-5124-front-panel.htm...
Pin 1, 5V fused ?
Regards, Stephan
(5,989 Views)
Yes , I want the max current out of pin 1 (+5V fused) of the mini DIN connector to supply a custom circuit.
The PFIs max current is specified with 24mA but not the 5V
Meanwhile I found the spec a spec of the 5102 (that don't have and use) in the high speed digitizer getting started, that notes(p22) that the 5102/5112/5911 is fused with 1.1A self resetting , 30mA recommended... however no information is provided for the 5922/5124. The pinout and ratings really should be noted in the specs 😉
So, do the 5922/5124 have the same rating than the 5102??
I want to drive a latching (bistable) relay to inject a signal. I added my planed schematic. Ontimes should be around 10ms, Coil resistance 173 Ohm, about once a min...
Don't know if D1 is needed , however it doesn't hurt 😉
Henrik
LV since v3.1
(5,980 Views)
Accepted by topic author Henrik_Volkers
Hi Hendrik,
according to my internal documents, the
NI PXI-5922 - National Instruments
https://www.ni.com/en-us/shop/model/pxi-5922.html
Order Number 779153-01
use a 1.1A polyfuse for the AUX +5V (fused) pin.
The
NI PXI-5124 - National Instruments
https://www.ni.com/en-us/support/model.pxi-5124.html
Oder Number 778757-01/02/03/04
use the same fuse with 1.1A rating.
Regards, Stephan
(5,974 Views)
Thank you 🙂
Henrik
LV since v3.1
(5,967 Views)