PXI
Connection of NI PXI 4110 tO TBX-68T
Hi
how do i connect the power supply to TBX-68T
pictures will help a lot
i need more than 20 Volts...........
Message 1 of 12
(5,360 Views)
Hi yongster,
The output of the 4110 can be used in various ways to achieve different voltages. There are a few things that need clarification:
- What voltage are you trying to output?
- What do you want to do with the power supply on the TBX-68T?
- Most DAQ card connected to the TBX-68T can only take a maximum of 10V - where are you routing the signals to?
Abhinav T.
Abhinav T.
Applications Engineering
National Instruments India
LabVIEW Introduction Course - Six Hours
Getting Started with NI-DAQmx
Measurement Fundamentals
Message 2 of 12
(5,348 Views)
Hi
great to hear from you
i have attached some slides to better explain the scenario, hope everything is clear if not let me know........
please give advice on connections
thks
Message 3 of 12
(5,337 Views)
Yongster,
Your powerpoint presentation is not very clear. Here's what I think you're trying to do:
- there are 8 connections
- 4 go to RTDs to measure temperature
- 4 control heaters
The RTDs can be connected to the TBX-68T by connecting IEX+ and IEX- channels (pins 12 and 45) to the excitation inputs of the RTD. The output of the RTD should be connected to the input channels on the TBX-68T.
In order to control your heaters, you can connect the power output on the 4110 to your heaters. In order to get more than 20V on the 4110, you will need to cascade the outputs in series.
Can you confirm this is the setup you are looking for?
Abhinav T.
Abhinav T.
Message 4 of 12
(5,318 Views)
hmmm, that sounds like the setting i need
in any case i have attached another attachment
please help to see if this is correct and if this is what you mean
thks
regards
Message 5 of 12
(5,300 Views)
Everything looks to be in order on this. I had one concern, what is the model number of your RTD? Is it a 2-wire, 3-wire or 4-wire RTD?
Abhinav
Abhinav T.
Message 6 of 12
(5,285 Views)
Thks
but to follow up i also have 2 concerns one related to the RTD
(1)RTD
for the RTD, we are actually using the properties of the fabricated metal using this equation (R = R0 (1 + TCR (T-T0)) which is integrated to  our chip as  a RTD, so for our case, can we use it as a 2 wire RTD?if so for the labview, we be reading data as temperature using RTD correct?
(2) For the heater, for the connections as shown previously, it seems that we are not able to control the individual heaters (i.e. the change in voltage supplied to the heater will be the same), i say if we want each heater individually to react to its respective temperature sensor (i.e. more fexibility, ), is it possible?
thks for the great support
Message 7 of 12
(5,276 Views)
1) Your RTD should be fine as it is connected - you can confirm this using this Developer Zone document.
2) In order to control the heaters individually you will have to use the three outputs of the power supply individually (or use more power supplies). If you try to control each heater using only one output of the power supply, all you heaters will be tied to that one voltage.
Abhinav
Abhinav T.
Message 8 of 12
(5,270 Views)
Nice post here, thanks for the info.
I am also planning to use for my thermal controller:
- PXI 4110 for heater drive.
- PXI 4351 for RTD interface (4-wires meas.!!), using the block terminal TBX-68T.
As I can read from your posts, one issue is unclear to my eyes:
- Do I need to supply excitation for my 4-wires measurement using PXI 4110?
Message 9 of 12
(4,492 Views)
I just forgot to add this, in case my question is not clear enough:
I see that there are current excitation channels in 4351, but:
I interpret from your drawings that you are wiring the heater drivers from 4110 through the 4351 terminal block TBX-68T. I thought this terminal block was intended to be used only for the 4351 module, for thermal readings, not for power output.
That is why I am asking.
Message 10 of 12
(4,485 Views)