PXI
DMM 4071 uncertainty for AC voltage measurements
I've found some good resources describing how to calculate the uncertainty of the 4071:
http://forums.ni.com/t5/PXI/DMM-4071-accuracy-4-5-digits/td-p/1962633
http://digital.ni.com/public.nsf/allkb/9FAB5BCA55D1E693862579F00076772D
Unfortunately none of them covers the uncertainty calculations for AC voltage measurements.
Anyone got a good resource?
So far I've got
  1. AC voltage accuracy, % of reading + % of range
  2. Temperature coefficient
  3. RMS noise, calculate aperture time and NPLC for my 5 1/2-digits resolution.
    I need to add value from graph  (on page 3 of specification) since the specs are for 6 1/2-digits, right?
  4. NO rms noise multiplier for AC measurements
  5. Common mode errors - How do I calculate these?
  6. NO crest factor errors - The algorithm used in 4071 is insensitive to crest factor.
Am I doing this right?
Any pointers are welcome.
(4,413 Views)
Hello JockeR,
I believe you are in the right direction!
Regarding your question for the common mode error:
If we take a look at any differential amplifiers course, we find the following formulas:
The first one is the CMRR as a ratio, the second in decibels and the third is the error.
CMRR is available in decibels in the datasheet:
The delta Vcm is the applied peak voltage.
Apply these in the last formula and you get the common mode error.
Bozhidar
(4,363 Views)