PXI
Error -1074135024 occurred at niFgen Initiate Generation.vi
I was calibrating a PXI 5124 with a NI-PXI-1042 when a error ocurred:
at step Verify Main Analog Path OffsetInvalid value for parameter or property.
Device: PXI1Slot4
Channel Name: 0
Recommended Property to Modify: Offset
Value: 1.750000
Maximum Value: 1.500000
Minimum Value: -1.500000
Conflicting Properties: Gain
Value: 3.0
Status Code: -200269
Possible reason(s):
IVI: (Hex 0xBFFA0010) Invalid value for parameter or property.
Calibration As Found As Left
Gain Test Value Low Limit Reading High Limit PassFail Low Limit Reading High Limit PassFail
12 60000, V 59485, V 59996, V 60515, V Passed 59785, V 59996, V 60215, V Passed
12 -60000, V -60515, V -59990, V -59485, V Passed -60215, V -59990, V -59785, V Passed
10 50000, V 49570, V 50001, V 50430, V Passed 49820, V 50001, V 50180, V Passed
10 -50000, V -50430, V -49991, V -49570, V Passed -50180, V -49991, V -49820, V Passed
7 35000, V 34697, V 35004, V 35303, V Passed 34872, V 35004, V 35128, V Passed
7 -35000, V -35303, V -34989, V -34697, V Passed -35128, V -34989, V -34872, V Passed
5 250000, V 247825, V 250045, V 252175, V Passed 249075, V 250045, V 250925, V Passed
5 -250000, V -252175, V -249890, V -247825, V Passed -250925, V -249890, V -249075, V Passed
3.5 175000, V 173463, V 000000, V 176537, V N/A 174337, V 000000, V 175663, V
N/A
What is the problem and how to fix it?
Thanks for the help.
(5,478 Views)
Hi Jhsilva,
Apparently, your offset value has to be beetwen -1.5 and  +1.5.
So, you have  to decrease your offset value.
Best regards
(5,450 Views)
Yes, i did understand that, but the question is how?
I cannot change the value of offset without actually change the Calibration Executive procedure correct?
Or is it possible to do that in some other way?
(5,447 Views)