PXI
Error in reading from niDmm: NaN
This error occurs often. It passed already but then fail again with NaN.
Do you also have example to capture the error and save to text file? Thanks
(4,828 Views)
Phuc,
Usually the DMM will return NaN when the measurement is out of range. If you increase the range of the measurement do you still get the error?
_
Mike
(4,827 Views)
Michael,
Thanks for the response. I will switch to auto-range.
Do you have any method to trap this error? when I get NaN error, I would like to set it to -99.99.
Best Regard,
PN
(4,827 Views)
Depending upon your software environment ther should be a function called "Is a number?" or "Not a Number?" In LabVIEW this can be found in the comparison menu. The return of the function will be a boolean and can use this to replace the number with -99 if needed.
Good Luck,
Mike
(4,827 Views)
The 'Not a Number' worked well.
After I set niDMM to autorange, I still got some NaN reading. Thanks to your recommendation, I set it to -99.9 now.
Phuc
(4,827 Views)
How to notify "NaN" in VB .NET after niDmm got this value?
(4,755 Views)
Hey everyone,
niDMM has a great function that you could call: (from help file)
[niDMM_IsOverRange
Usage:
ViStatus = niDMM_IsOverRange(ViSession Instrument_Handle, ViReal64 Measurement_Value, ViBoolean* Is_Over_Range)
"Takes a measurement value and determines if the value is a valid measurement or a value indicating that an overrange condition occurred."]
You could call this after a read and check for the over range condition. Immediately following you could check the value of the ViBoolean that is set and return your value accordingly. i.e.
//sample
ViStatus = niDMM_Read(...................., Reading)
ViStatus = niDMM_IsOverRange(........, Reading , Is_Over_Range)
if(Is_Over_Range)
Reading = -99.99;
//rest of code
Just a thought
Sincerely,
Gavin Goodrich
Applications Engineer
(4,743 Views)