PXI
How to solve Error 250043 when synchronizing pxie 5423 and 5105
When I synchronize PXIe-5423 and PXIe-5105 with NI-TClk, an error 250043 usually occurred, and the error details are:
Error 250043 usually occurredat niTClk configure For Homogeneous Triggers. vi: 201001.
Possible reasons:
NI-TClk invoked a call on a session, and the instrument driver reported an error.
Error reported by the instrument driver:
IVI:(Hex 0xBFFA0010) Invalid value for parameter or property.
Session index (starts at zero): 1.
The driver versions are: scope 18.6 and fgen 18.1.
Could you give some advice to solve this problem?
(2,845 Views)
Error -250043 occurs at niTClk Synchronize when synchronizing different device models
An error can occur if the TClk timebase of the Sync Pulse Sender device is not a multiple of the timebase of each slave device.
Workaround: If you have set the Absolute Sample Clock Offset or Phase DAC Oscillator NI-SCOPE properties on synchronized devices of different model numbers, place a different device model first in the array of session references that are being synchronized.
More information: https://www.ni.com/en/support/documentation/bugs/18/ni-scope-18-5-known-issues.html#705654_by_Date
Liu Sang