07-15-2014 03:47 PM
Hi, I need to characterize phase and amplitude imbalance of two quadrature signals from a 14bit TxDAC. I selected two 5162 modules to do the job. The signals can have a frequency of up to 1.2GHz..
Is PXIe-5162 capable of doing this job?
The plan was to record a long sample of the signals and then do the processing in ie. Matlab.
I am facing a few problems. Is it possible to synchronize sampling of all channels on one 5162 module, and how to synchronize sampling of 2 x 5162 modules with enough accuracy to have a phase resolution below 1deg?
Are there any examples of how to synchronize sampling of multiple channels/modules in LabView ?
Can the phase and amplitude measurement be done directly in LabView without involving Matlab?
07-17-2014
09:17 AM
- last edited on
08-06-2024
04:09 PM
by
Content Cleaner
Hi Robert,
Did you already read some white paper and KB about synchronization using LabVIEW?
https://www.ni.com/en/support/documentation/supplemental/10/synchronization-explained.html
You can find some other white papers on ni.com. I think this will be a good start before doing something else.
You also can search in the LabVIEW examples.
Here some information about the measurement of the phase and the amplitude:
Let me know if it is good 🙂
07-17-2014
02:11 PM
- last edited on
08-06-2024
04:09 PM
by
Content Cleaner
Thanks for a reply.
I read a bit about the NI-TCLK synchronization. Does it apply to module-to-module sunchronization or is it used within one module too?
I found this page
Overcome the Challenges of Building High-Channel-Count, High Speed Measurement Systems
It says 30ps skew after manual calibration, which is not enough for my application.
I didn't find any information about the sampling skew for converters within one module.
I recall a number 12ps from some previous research but I cannot find it now. 12ps within one module is still too much.
07-18-2014
08:25 AM
- last edited on
08-06-2024
04:11 PM
by
Content Cleaner
Hi Robert,
From the datasheet of the 5162 (https://www.ni.com/docs/en-US/bundle/pxie-5162-specs/page/specs.html) page 20, its says that using the manual calibration you can synchronize two identical 5162 with less than 5ps.
Here some other information:
Is that answering your questions?
Have a nice weekend.