PXI

cancel
Showing results for 
Search instead for 
Did you mean: 

PXI 4130 Power SMU Output

Hi, I am trying to get my head around how the PXI 4130 works. I have a solenoid valve that I am trying to actuate with the utility channel of the 4130. It says that the utility channel can source 6V and 1A. The solenoid valve requires 6V, but the coil in the valve says it is a 12 Watt rated coil. Since amps = watt/voltage, the solenoid valve should draw 2A (12/6). But does that mean that the 4130 will not let out 2A and will only let out 1A? If I am supplying 6V will it not automatically just output 2A by the law V = IR??

 

 

0 Kudos
Message 1 of 4
(6,318 Views)

Your understanding is correct.  The 6V Utility Channel (Channel 0) can only output 1A maximum on the PXI-4130 SMU.

 

If your load tries to draw more than the current limit (user programmable up to 1A on Channel 0), then the SMU output will go into "current compliance" which forces the SMU into constant current mode instead of constant voltage mode - as a result, your voltage will drop accordingly until V=IR.  This behavior is often used to protect the device under test connected to the SMU.  ...I might start by testing out the solenoid and seeing if it will actuate at only 1A, but the best solution would be to use an output capable of 2A like Channel 1 on the PXI-4130. 

 

Also, please take note that heavily inductive loads can occasionally cause issues with the control loops on power supplies on SMUs:

http://zone.ni.com/reference/en-XX/help/370736F-01/ni_dc_power_supplies_help/4130_source_stability/

NI has recently developed some cool technology on newer SMUs that helps manage control loops under reactive loads:

http://www.ni.com/white-paper/13093/en/

 

For both points above, I might recommend to monitor what happens to the 4130 output as you actuate the solenoid the first time - either using the on-board voltage and current measurements of the 4130 (can read at up to 3kHz) or with an external scope on the voltage.

0 Kudos
Message 2 of 4
(6,312 Views)

yea i am using the 4130 slot for something else tho.. I actually have 2 of the 4130 slots, do you know if I can take out one of the other slots in my PXI box (such as 2564) and and swap it with my second 4130 slot and have two of them? is it as easy as that?

0 Kudos
Message 3 of 4
(6,306 Views)

No problem with that. if your chassis has space, and the slot is compatible there is no problem. Remember you need to power off the computer and the chassis to make changes.

 

Regards

Frank R.