I am trying to use a PXIe-4080 to measure thermocouple level signals to validate a sensing circuit (sensor). I have a thermocouple simulator to provide the rough voltage levels, and then I want to compare the values measured by the PXIe-4080 and my sensor. When I have my sensor hooked up to the simulator it reads the correct value (within a few uV) and has low noise. When I connect the negative probe of the PXIe-4080 to the negative terminal of the sensor the reading shifts by ~10uV and a lot of noise shows up (~+/- 20uV). This does not go away when I connect the positive probe. This does not seem normal, since the PXIe-4080 probes should be isolated and measuring differentially.
I don't believe this has anything to do with my specific sensor because if I use an Agilent 34401A DMM I do not see this effect.
I am using the Soft Front Panel, I have tried 10MOhm and 10GOhm input impedance to no avail.
Any ideas of what to check or what could be wrong?
Solved! Go to Solution.
08-09-2018
09:33 AM
- last edited on
05-08-2025
10:07 AM
by
Content Cleaner
Hi RyanMi,
Here are some ideas:
You could check Table 1 of the following document: https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider...
And check this other two:
https://www.ni.com/docs/en-US/bundle/ni-dmm/page/noise.html
https://www.ni.com/docs/en-US/bundle/ni-dmm/page/dc-noise-rejection.html
Hope this helps!
08-17-2018 09:09 AM
Hi A-Guz-Man,
Thanks for the suggestions. Unfortunately the noise/offset is well suppressed in the PXIe-4080 output but is induced into our sensor output when we attach the cables from the PXIe-4080 (but not the Agilent 34410a)
What we eventually ended up trying is putting a capacitor across the lines and that seemed to clear up the issue.