PXI
PXIe-4080 Noise and Offset when measuring thermocouple
I am trying to use a PXIe-4080 to measure thermocouple level signals to validate a sensing circuit (sensor). I have a thermocouple simulator to provide the rough voltage levels, and then I want to compare the values measured by the PXIe-4080 and my sensor. When I have my sensor hooked up to the simulator it reads the correct value (within a few uV) and has low noise. When I connect the negative probe of the PXIe-4080 to the negative terminal of the sensor the reading shifts by ~10uV and a lot of noise shows up (~+/- 20uV). This does not go away when I connect the positive probe. This does not seem normal, since the PXIe-4080 probes should be isolated and measuring differentially.
I don't believe this has anything to do with my specific sensor because if I use an Agilent 34401A DMM I do not see this effect.
I am using the Soft Front Panel, I have tried 10MOhm and 10GOhm input impedance to no avail.
Any ideas of what to check or what could be wrong?
(2,846 Views)
(2,819 Views)
Accepted by topic author RyanMi
Hi A-Guz-Man,
Thanks for the suggestions. Unfortunately the noise/offset is well suppressed in the PXIe-4080 output but is induced into our sensor output when we attach the cables from the PXIe-4080 (but not the Agilent 34410a)
What we eventually ended up trying is putting a capacitor across the lines and that seemed to clear up the issue.
(2,803 Views)