PXI
PXIe-6570 Force/Measure Current Accuracy Computation
I am having problems on understanding the computation for PXIe-6570 Force/Measure Current Accuracy.
What comes next after finding the zone (step 6)?
Can anyone give me an example of the computation?
Thanks!
(1,448 Views)
Accepted by topic author trixx25
Let me show you an example for calculating accuracy,
Steps
  1. FI = 1.5mA
  2. Using 2mA as the best range
  3. % Force current range = (1.5m / 2m)*100 = +75%
  4. Assume my impedance is 1k Ohm, Voltage Required = 1k * 1.5m = 1.5V
  5. Plotting the 1.5V, +75% in the Figure3, the point lies in the grey area which is designated as zone 1
  6. Looking at Table1, it states that +-1% for Zone1 irrespective of range
  7. Though not explicitly stated, you calculate the accuracy as you typically do for any measurement or value in engineering, apply +-1% to +-2mA range, now it becomes +-40uA because +-2mA constitutes 4mA total range.
Hope step 7 clarifies your question about how to apply +-1% to get the accuracy number.
Alternatively, if my impedance was 100 ohm and force current was -1.5mA, the voltage required becomes 150mV. Now, the 150mV, -75% lies outside zone1 in the graph and hence there is no warranted accuracy specification (as mentioned in step5)
Santhosh
Soliton Technologies
(1,414 Views)
It makes so much more sense now.
Thank you!
(1,387 Views)
