vi_attr_tmo_value is "Invalid Property"

Hi,

 

I am running a data acquisition system using a PXI1045 chassis (~from about 2006).  Recently I noticed that the power button is solid red.  The PXI manual explains what a green & flashing red LED mean, but not a solid red LED. 

 

In MAX, I opened the VISA Test Panel and it says that there is an error.  The Error says that (picture attached):

   Attribute:  Timeout

   VISA Attribute Name:  vi_attr_tmo_value

   Current Value:  Invalid Property

 

The code was recently converted from LabView 2006 to LabView 2013.  The program seems to be running ok, though it is a rather large program and hasn't been thouroughly reviewed yet.

 

Any thoughts on what the issue might be, and if I should be concerned?

 

Thanks,

Eric

 

 

Message 1 of 2
(6,520 Views)

If the power button ont he chassis is solid red, that means the exhaust fans on the chassis may not be working properly:

 

http://digital.ni.com/public.nsf/allkb/A15DFD6915FED85386256E680061D55D

 

You will want to RMA this Chassis.

 

P.S.  Kudos for searching the manual before posting to the Forum!