06-18-2009 07:10 AM
In my test system, I need to use PXIe-5673 for the GSM/EDGE/WCDMA modulation application. I also need to test the IIP3(Two-Tone Third-Order Intermodulation Distortion Measurement) performance. As I can see from this link:http://zone.ni.com/devzone/cda/tut/p/id/4384, It states that two signal generator(PXI565X) are needed for the IIP3 measurement. As I already have a PXIe-5673, Is it possible to test IIP3 using one single PXIe-5673 ? or I need to buy one more PXI565X, that means PXI565X+PXIe-5673 for the IIP3 measurement? I am really new guys to the RF field application. Highly appreciate to any kindly reply!
06-23-2009 04:50 AM
06-23-2009 05:28 AM
The reason I raise this question is that: the NI PXIe-5673 is consist of three parts:
1. NI PXIe-5611RF upconverter
2. NI PXI-5652 RF continuous wave (CW) source (carrier wave)
3. NI PXIe-5450 dual-channel arbitrary waveform generator (AWG).
Is it possible that we just generate two-tone signal by the NI PXIe-5450 ,then upconverter these two tone signal by the upconverter PXIe-5611RF ?
06-23-2009 06:51 AM
06-23-2009 11:42 PM
Hi Jerry,
Thank you very much for your reply.We are trying to introduce PXI solustion for semiconductor test for the production test. If the PXIe-5673 is capable of testing the IIP3, we would not willing to add one more signal generator.Because,you know,the test cost is very sensitive for the production test. The following is the test specification of DUT. Would you help to have a look?
| Parameter | Condition | Frequency | Min. | Typ. | Max. | Unit |
| Intermodulation intercept point (IP3) | Two-tones 900 MHz, 5 MHz spacing +10 dBm each tone | 0.5–3 GHz | 46 dBm |
As I can see from the datasheet of 5673:", NI PXIe-5450 is capable of producing more than 100 MHz of RF bandwidth". For the frequecy,I think it should be able to generate two-tones 900MHz with 5MHz spacing.Is it? But I am not quite sure about the Intermodulation Distortion (IMD3) of the PXIe-5673's signal if we generate it in this case. Do you have any information I can refer to?
It will be highly appreciated if you can give some suggestions on it.
07-01-2009 04:01 AM
Hi Jerry_L,
It's me again. I do lot of research these days. Would you mind spend some time to see if my understanding is correct or not?
As I see from the user manual of PXIe-5673, the worst IMD3 is –48 dbc while the two tone spacing 300KHz.
But base on the IP3 calculation formula: "IP3=Pout+|IMD3/ 2| "and the product specification(see my reply above) of my DUT, the DUT's IMD3 performance should as below:
IMD3(DUT)=-2*(IP3-Pout)=-2*(46dbm-10dbm)=-72dbc.
That means the IMD3 performance of PXIe-5673(-48dbc) is worst than the DUT's (-72dbc). So PXIe-5673 is not suitable for my application.
But I still have concern is that:
These IMD3 are measured under different condition.The PXIe-5673 is measured with two tones, 300 kHz apart, at –6 dBm per tone; And the DUT specification is measured with Two-tones 900 MHz, 5 MHz spacing +10 dBm each tone
Does the IMD3 performance vary a lot when changing the spacing width?Do you have the IMD3 performance of PXIe-5673 with 5MHz spacing? Appreciate it if you can give some suggestions