Guys,
I have a PXI-5661 RF VA and PXI-5142 connected together to gather data generated from a PXI-5650. Using the RFSA soft panel and Measurement and Automation Explorer I cannot get the signal analyzer to see the signal generator.
I am trying to perform a simple loop back test to verify the hardware is connected correctly and functioning.
Hardware Set Up:
PXI-5661 RF In is connected to the PXI-5650 through a 6dB attenuator.
PXI-5661 10MHz out(1) is tied to the PXI-5661 PXI 10MHz I/O with the supplied cable
PXI-5661 10MHz out(2) is tied to the PXI-5142 CLK In with the supplied cable.
PXI-5661 Output is connected to the PXI-5142 CH 0 input.
I have associated the down converter and the IF digitizer as described in the getting started guide.
Using the defaults, the center freq for in/output are set to 1GHz. I am not seeing a change in the incoming RF signal when I change the center freq or the output power level.
Should I connect the CLK OUT of the PXI-5142 to the Ref In terminal of the PXI-5650?
Any help would be greatly appreciated. Thanks in advance. If you require additional information, please let me know.
Thanks,
-Joe
07-29-2010 02:35 PM
Hello Joe,
The hardware set-up is correct.
Do you have a different signal generator and can you use that to output a signal into the 5661? Can you see a signal there?
Do you have a different spectrum analyzer and can you use that to analyze the signal being generated by the 5650?
Finally can you show us the screen-shots of the MAX Test panels for your RF devices?
Thanks