RF Measurement Devices
VST and RFSA Soft Panel
Hi all,
I want to use the NI RFSA Soft Front Panel with my VST PXIe-5646R, but I always get the error:
Error: The Maximum Time was exceeded before the operation completed.
Is this normal?
I have changed the settings (ex: Span to 1MHz), went back to the preset, but I always get the same error.
I have tried the RFSG Soft Front Panel, and it works fine, I can generate a CW correctly.
Is there anything special I need to configure in order to use the NI RFSA Soft Front Panel with VST 5646R?
System:
PXIe-1082Q Chassis
PXIe-8135 controller
VST 5646R in slot #2
Conduant Drive in slot #5
Windows 7 x64
NI-RFSA 14.0
Thanks in advance,
Serge
serge.malo@skydelsolutions.com
(6,008 Views)
There shouldn't be anything special to do in order to control the 5646R.  What are the steps you're doing after the preset button?
I was able to set span to 1 MHz after clicking preset without error.  I am using RFSA 14.1, but that shouldn't make a difference in this case.
(6,003 Views)
Hi Serge,
Could you try running niVST Self Calibrate.exe in C:\Program Files (x86)\National Instruments\NI VST\Self Calibration (or the equivalent path on your machine)? That may give us a better indication of what is going on.
Best Regards,
Peter W.
RF PSE
(5,997 Views)
Thanks for the quick answer,
After "preset", I only  let the Soft Front Panel run until the error shows up. I set the span to 1MHz, and got the same error.
I tried the Self-Calibration and it failed. Here is the error message:
Calibration of the device failed.
Calibration failed for device RIO0 for the following reason:
Error -366011 occurred at an unidentified location
Possible reason(s):
A local oscillator PLL lock error occurred. If the Frequency Settling Units property is set to Time After I/O, make sure that the Frequency Settling property is set to a value sufficient for this frequency change.
If you are using an external reference, make sure that it is connected and meets the frequency accuracy and voltage level requirements.
I tried to find the "Frequency Settling Units" property, but I did not see it with the VISA Test Panel.
Does this error tells you something?
Serge
(5,991 Views)
Hi again,
In your opinon, is the unit broken? Should I open a case with NI support?
Serge
(5,953 Views)
Hi Serge,
Yes please reach out to the applications engineers directly. They may want to troubleshoot a bit more before bringing the module in for RMA but there is potential that it will be necessary.
Best Regards,
Peter W.
RF PSE
(5,948 Views)
Hi all,
Just to close the loop:
I had a discussion with a NI Application Engineer on the support line today. He suggested testing the VST with different combinasons of
-Slot number
-Chassis
-Controller
The Self-Calibration failed in all combinasons, so he suggested to send the unit back for RMA.
Thanks for your help,
Serge