Configuring IO delay in NI 6587 for Data Acquisition
I am trying to read LVDS data using 6587 adapter module with 7966R FPGA module. The external clock I receive looks identical to data (just hi-lo toggling). In order to capture the data I need to shift the phase of the clock by a quarter clock cycle to get the most stable data values. Since the clock IDELAY is not accessible (always set to 0) I am thinking of shifting all my data accordingly. I have some questions on how to use DATA_IDELAY_INCREMENT for this purpose.
a) Logic high level on DATA_IDELAY_INCREMENT increases data delay by one tap per ACQ_REGIONAL_CLOCK, where one tap is 78.125ps nominally, so what should I do to get a delay of  400ps ? Is it related to number of rising edges on DATA_IDELAY_INCREMENT ?
b) How can I use IDELAY_CALIBRATION_CLOCK ?
(5,061 Views)
Accepted by topic author mbothra
mbothra,
To answer your first question, you would need to have a configuration loop that sets the DATA_IDELAY_INCREMENT to true for as many clock ticks as you need to reach 400 ps. The DATA_IDELAY_INCREMENT node sets the enable input of the IODELAY Xilinx primitive to true, and the direction of the IODELAY is hard-coded to always increment. What this means is that for every clock tick where the DATA_IDELAY_INCREMENT node is true, the delay will increase by 78.125 ps.
In your case, you would need 400/78.125 clock ticks, or 5.12 (rounded either down to 5 clock ticks for a delay of 390.625 ps or up to 6 clock ticks for a delay of 468.75 ps). To implement this, you could have a single cycle timed loop run for the number of iterations you need, stop the loop, then start your actual data acquisition loops.
For your second question, the IDELAY_CALIBRATION_CLOCK is locked at 200 MHz. There are no changes that you can make to this value to cause any kind of delay.
(5,028 Views)
Thanks !! Smiley Happy
(5,023 Views)