NI 9213 Calibration Verification
From the procedure NI supplied (attached), it looks like all the ‘set up’ portion of the procedure can be done using the ‘DAQ Assist’ module by editing the different settings, is that correct? Or do I need to draw out the 'set up' as directed in the Calibration Manual.
Is there a sample time read different between "High Resolution" and "High Speed" in the NI9213?
For verification of the NI 9213 channels, I am using the 'DAQ Assist':  inside 'DAQ Assist, Advance timing' I had to change the 'Additional Timing Settings' to accommodate the number of Samples I wanted to read to prevent an error.
In the calibration document, on Page 8, Step 7, is it true that in 'High Speed', the Read rate is much faster? But when I take 50 sample in "High Resolution" and "High Speed", the time it takes the DAQ to perform the read is the same? This is all done assuming I can use the "DAQ Assist" software module.
(5,160 Views)
Hi AATEE, it is recommended to use the pdf guide you already have, therefore not using the DAQ assistant but the separate VIs for DAQmx tasks.
Eric NI