We have a new NI-DAQmx and two NI 9213 TC input modules. Need to monitor Temps on several active devices. In this AC-to-DC conversion module there could be voltages on the order of up to 60VAC or 150VDC on the Drain of a MOSFET. Would physically attaching Thermocouples to these live voltages cause any issue at the NI 9213 input, or do they need to be electrically isolated?
Solved! Go to Solution.
07-01-2010 04:52 PM
Hi Bob,
The overprotection voltage for that module is listed at 30 V between any two inputs. Voltages in the range you mention have the potential to damage the card. You’ll have to electrically isolate those voltages from your NI 9213 module.