Hi Steve,
The offset you are getting from the calibration node is NOT for offsetting the
actual measurement but rather this offset is used for converting the binary
data to nominal (meaningful) data.
For more information about how calibration works please search for the following in the LabVIEW help:
“NI 9237”
“Converting and Calibrating CompactRIO Analog Input Values”
Additionally, please take a look at the following example about offsetting your actual measurement (zero-out an unloaded bridge)
http://zone.ni.com/devzone/cda/epd/p/id/5816
Lastly, in the LabVIEW example finder (Help » Find Examples…), browse to Hardware Input and Output » CompactRIO » Module Specific » Analog Input » NI 9237 Getting Started.lvproj
06-06-2008 02:17 PM
Eli,
Thanks for the response. However, I'm using the calibrated values (Fixed Point) output of the 9237 not the binary values. My question is more a hardware question in wanting to verify exactly what the module is doing when you switch the Offset calibration to True. I thought I had read somewhere that the only thing the module does is disconnect the AI lines from the bridge and short them together. Is this true?