I am using USB 4065 multi-meter to measure the electrical resistance of the my sample (2-wire). I wonder if it is possible to change the reading rate in Instrument Studio? Right now it is about 5 data points per second, I would like to increase it to 20 points, or as close as possible.
Any help will be appreciated.
Thanks
05-21-2019 09:41 AM
Have you tried going into channel settings and setting a resistance range, then changing the resolution digits? The device will be able to capture faster at fewer digits of resolution. Will your use case support this?
05-21-2019 02:42 PM
Yes, I did. Changing a resistance range and resolution digits gave me up to 10 data points per second. I wonder if there is a way to modify it quantitatively?