Hi Zyl99,
It looks like you may have created a ground loop in your configuration. Your signal source is most likely a grounded source, and the 9201 only supports referenced single ended (RSE) connections. For more information on ground loops, refer to this DevZone article (your connection setup probably looks like the 'Not Recommended' configuration in Table 1):
http://zone.ni.com/devzone/cda/tut/p/id/3394
You will need another device (such as the 9215) that can accept differential or non-referenced single ended (NRSE) connections to measure grounded signal sources.
Let me know if you have any other questions.
Regards,
Way S.
NI UK Applications Engineer