Hi there

 

I am using DAQ 6215 (USB) to perform step response of a device, i am using transition and pulse measurement function to determine the rise time and fall time 

but it doesnt seem to work for the frequencies below 0.5hz how can i solve this issue is there any other way and also how can i determine the settling time of the response signal please suggest me something i am attaching my VI just check it out thank you........

0 Kudos
Message 1 of 8
(3,996 Views)
You seem to be missing two of your attached VI's, the ones that configure your tasks so I can't tell how your timing is setup.  Are you getting an error occuring?  I was able to perform a rising edge time analysis at 0.25 Hz without issue so long as I was taking a large enough chunk of the waveform so that the analysis VI could see the rising edge.
Doug Farrell
Solutions Marketing - Automotive
National Instruments

National Instruments Automotive Solutions
0 Kudos
Message 2 of 8
(3,978 Views)

Hi Doug,

Sorry for missing out the sub VI's here i am attaching them.....thank you for the reply 

Download All
0 Kudos
Message 3 of 8
(3,972 Views)

Hi Doug,

 

I have set the sampling rate (o/p) and samples to write to 800000 similarly sampling rate (i/p) and samples to read to 250000 it (loopback) seemed to work i.e it fast and the transition measurement block doesnt seem to bother me at low frequencies but i still have to test it real time i dont have the device now once i do the test i will let you know if it is really working thank you......bye.....

0 Kudos
Message 4 of 8
(3,969 Views)

Hi Doug,

 

With the above mentioned settings there is another problem at frequency 0.1hz the frequency i detected using extract single tone measurement indicates 99.7k and it fluctuates between this value and 2.07its not constant it is the same case till 2hz i dont understand the problem pls check this i am attaching my VI with slight changes thank you bye......

0 Kudos
Message 5 of 8
(3,967 Views)

The problem is not with the VI, it is with the timing parameters you are using.  You are acquiring 250,000 samples at a time at 250,000 samples/second.  Meaning that you are acquiring (and analyzing) 1 second of data at a time.

 

To be able to perform a single tone extraction properly you need at least double the period of the waveform, this is why the issue resolves itself at 2 Hz (or with a period of 0.5 seconds).  If you are acquiring and analyzing less waveform data than this, you will not read the correct frequency as there will not be enough data.

Doug Farrell
Solutions Marketing - Automotive
National Instruments

National Instruments Automotive Solutions
0 Kudos
Message 6 of 8
(3,948 Views)

Hi Doug,

 

 I am able to determine the overshoot, rise time and fall time but how can i determine the settling time of the response signal can you give me a suggestion bye.........

0 Kudos
Message 7 of 8
(3,912 Views)

Settling time is generally accepted to be within 2% of the desired amplitude.  There is no VI or anything that I can find to do this in LabVIEW.  What I would recommend is that you have to just do a manual compare of the signal you are reading to the desired amplitude and flag the time at which it gets within 2%.

 

http://en.wikipedia.org/wiki/Settling_time

Doug Farrell
Solutions Marketing - Automotive
National Instruments

National Instruments Automotive Solutions
0 Kudos
Message 8 of 8