Hi Jon,
The problem is that NI seems to recommend (in tutorials etc.) the use of RTI voltage errors to calculate measurand errors (strain, temperature etc.), and as you mentioned this becomes unfeasible as the output voltage from the sensor goes to zero. It would make more sense if absolute errors were used instead. For example, if no strain was applied to the strain gauge in my application and thus its output voltage was 0V (ideally), the absolute voltage accuracy of +/-20uV could mean that the strain is reported as +/-16ue: is this the correct interpretation? I understand that the quoted voltage accuracies are usually conservative, but are possible. The absolute voltage accuracy in my case doesn't vary much over the measurement range as the gauge's output voltage is small, so the absolute strain accuracy is fairly consistent. Therefore the strain error as a percentage of the measurement will decrease for increasing strains.
Thanks for your continuing help.
Regards,
Brian