My current application is configured to monitor (measure) RTDs. I have an 1320 terminal block connected to an SCXI-1121 module. This interface is used to measure thin film RTDs using the two-wire method. My application has worked for several years using a certain type of RTD that has a resistance of 500 ohms at 0 C and ~ 820 ohms at ambient temperature (22 C). However, I am now trying to interface with a different model RTD. This model is of the same type (thin film with two wire application) but its response slope is different. It's 0 C resistance is ~1100 ohms and its resistance at 22 C is ~ 1800 ohms. When I connect this type of RTD and change my equation to account for the difference in response, my readings get maxed out at 60 C. I used a potentiometer to test the old set-up and it appears the readings max out at approximately 1.2 kohms. The new RTDs resistance appears to high for the set-up to read it. I am using current excitation of 0.15 mA. I thought that maybe the excitation was too low and reconfigured for an excitation of 0.45 mA. However, this did not improve the situation. I was still unable to measure any change of resistance with the higher impedence RTD. What could I possilbly try next with the configuration? I am absolutely sure that my linear equations are correct for both RTDs and that I have properly applied them through our software application.
Thanks,
James