First of all, I think the measurement hardware-section is a better forum for this question.
I don't think the TBX-68T is the problem. That are just some electronics parts to get rid of the AC-hum.
I urge you to take a look at the specifications of the hi-res DVM. If I interpret the data, it says that for a 60 Hz rejection, you have a maximum of 9.7 readings per second.
That means that you have new samples every 15/9.7 = 1.5 seconds, so you cannot get the 1 sec. speed you need.
Why is that?
Sigma-delta ADC's (the ADC of the 4351) are capable of hi resolution, but they use digital filtering-techniques inside the silicon to acchieve that.
And as you know, every filter (analog or digital) takes some time to settle to a stable read-out.
And exactly
that is the reason why a single-channel read-out is so much faster than a multi-channel read-out in sigma-delta designs. It doesn't have to wait until the output is settled.
So what happens, you try to read-out the DVM, but internally it waits for an end-of-conversion of the ADC. Your vi just waits until it is ready.
Note: I have not used the 4351 in practice but my company designed a compact digitiser with sigma-delta convertor under LabVIEW control.
Unfortuately it never reached the market, but it was nice to see temperature vary in the 0,001K range.
Hope this helps,
Patrick de Boevere
Serenity systems design