06-08-2006 12:26 AM
06-08-2006 04:35 PM
Hi Haider,
Are you using Traditional NI-DAQ (Legacy) or NI-DAQmx for this application? I think that you need to reset your devices to eliminate the error.
If you are just using NI-DAQmx, go to MAX and right click on the DAQ board and select "reset device." Do the same for the SCXI Chassis.
If you have been using Traditional NI-DAQ (Legacy), then you may need to reset the Traditional NI-DAQ driver by right clicking on "Traditional NI-DAQ (Legacy) Devices" and selecting "Reset driver for Traditional NI-DAQ." I hope that you find this information helpful.
Regards,
Hal L.
06-09-2006 03:32 AM
I have done "Reset device " on DAQmx. Still It is not detecting the modules from Chassis (1000DC).When I tried to reset the Chassis after reseting the device it signals an error"-88705 The Chassis reset has failed.The device may have not installed and so much/........."
Actually I am using
LabVIEW 7 EXPRESS
NI DAQmx 7.4
NI 6110 Card.
SCXI 1000DC +SCB 68 cable+module 1121,1124+terminal block+1320+1327
So what should I do?
06-09-2006 05:02 PM
06-16-2006 01:15 AM
dear sir
I have tried to download the DAQmx 8.1 four times through the link that you provided but after 7 or 8 hrs it only downloads to 509MB or less and refuse to unzip.
06-19-2006 10:20 AM